    <div class="moz-cite-prefix">Hi Reza,<br>
      <br>
      Thanks for your feedback. This is a good question, but with no
      easy good answer!<br>
      <br>
      For an arbitrary PET scanner, I think your suggestion of using the
      crystal width is not too far from the reality (considering no
      depth-of-interaction measurement).<br>
      <br>
      About the images you sent, there clearly are artifacts with 1mm
      FWHM which cannot be seen with 2mm FWHM. They can come from
      several sources:<br>
       - reconstruction of simulated data without proper normalization
      correction factors (in that case, the use of a PSF may
      artificially reduce such artifacts)<br>
       - a strictly line projector with too small voxels (if you are
      using Siddon for example; in this case try the Joseph projector
      which makes use of interpolations, it can decrease such artifacts)<br>
       - or a combination of the two.<br>
      <br>
      Cheers,<br>
      Simon<br>
      <br>
      Le 28/04/2017 à 18:55, M.R Teimoori a écrit :<br>
    </div>
    <blockquote
cite="mid:CACEZ9i_yFUdv5rEGFt1jN4K1zVeE=qhOmR=NLvg7VykD18F4Zw@mail.gmail.com"
      type="cite">
      <meta http-equiv="Content-Type" content="text/html; charset=utf-8">
      <div dir="ltr">Hello CASToR Users,
        <div><br>
        </div>
        <div>I've been using CASToR to reconstruct images of various
          human PET scanners to compare their image quality. So far,
          through working with CASToR, I learned that the quality of the
          reconstructed images for any PET scanner significantly varies
          when changing the image convolver's FWHM. </div>
        <div><br>
          For example, the attached file shows images of a Derenzo-like
          phantom obtained from simulations of a brain-dedicated PET
          scanner and reconstructed with CASToR using listmode EM
          reconstruction. This arbitrary PET scanner had LSO detectors
          with a crystal width of 2 x 2 mm^2 and a crystal depth of 20
          mm. In (A), images were reconstructed using "-conv
          gaussian,2,2,3::psf" and in (B), the same listmode data were
          reconstructed using the same reconstruction parameters but
          with convolution parameters of "-conv gaussian,1,1,3::psf". </div>
        <div><br>
        </div>
        <div>How legitimate does this sound to say that "for a PET
          scanner, the convolver's FWHM should be equal to the scanner's
          crystal width"? In general, are there any recommendations as
          to how image convolver's parameters should be chosen for any
          arbitrary PET scanner?</div>
        <div><br>
        </div>
        <div>Thank you for your responses,</div>
        <div><br>
        </div>
        <div>Cheers,</div>
        <div>Reza</div>
      </div>
      <br>
      <fieldset class="mimeAttachmentHeader"></fieldset>
      <br>
      <pre wrap="">_______________________________________________
Castor-users mailing list
<a class="moz-txt-link-abbreviated" href="mailto:Castor-users@lists.castor-project.org">Castor-users@lists.castor-project.org</a>
<a class="moz-txt-link-freetext" href="http://lists.castor-project.org/listinfo/castor-users">http://lists.castor-project.org/listinfo/castor-users</a>
</pre>
    </blockquote>
    <br>
  </body>
</html>