<p>Hello Maxime,</p>
<p>Thank you for your feedback, the current implementation of the
Siddon multi-lines projector class (and the associated <i>GetRdmPositionsAndOrientations()</i>
function from the iScannerPET.cc file) has indeed some
limitations. It is still work in progress and is only intended to
be used with usual systems whose detectors have an axial
orientation equal to 0. I should have made it clearer in the
description of the projector. <br>
It has not been assessed with systems using detectors with
non-zero axial angle, and as you noticed it contains some errors
in this case. Regarding the comment about <i>mp_meanDepthOfInteraction</i>,
it refers to a previous implementation in the event a DOI was
provided.<br>
</p>
<p>This projector is actually a draft and we plan to revamp the
management of multi-line projector to make it more flexible. In
short, the estimation of the random position in the detector
should be computed in the mother class (or inside an additional
intermediate class), from which any line projector class could be
called to compute the system matrix elements. This would avoid to
create a multi-lines projector class specific to each projection
algorithm. Besides, we plan to rework on the orientations in order
to make it more flexible for less generic systems. All of this
will require some substantial work though.<br>
</p>
<p>Thank again for your remark and suggestions. I will have a look
this summer to correct the current implementation. <br>
</p>
<p>Best,<br>
Thibaut<br>
</p>
<p><br>
</p>
<br>
<div class="moz-cite-prefix">On 18/06/2018 20:13, Maxime Toussaint
wrote:<br>
</div>
<blockquote
cite="mid:a7e0e131-5dcf-6c8c-bd1a-0e81add568a0@usherbrooke.ca"
type="cite">Hi,
<br>
<br>
I think there is an error in the GetRdmPositionsAndOrientations
function of the file iScannerPET.cc. In short, the vectors used to
create a random position in a detector are not orthogonal between
each other and two of them are not guaranteed to have a norm of
one.
<br>
<br>
The long version:
<br>
mp_crystalOrientation gives us a vector that goes along the depth
of the current detector. However, CASToR does not keep a similar
vector for the detector's transaxial and axial axes. For
simplicity, let's define (mp_crystalOrientationX,
mp_crystalOrientationY, mp_crystalOrientationZ) as (uX, uY, uZ).
Currently, the transaxial axis of the detector is defined as (in
the function GetRdmPositionsAndOrientations):
<br>
(uY, uX, 0)
<br>
while the axial axis of the detector is defined as:
<br>
(uX*uZ, uY*uZ, sqrt(1-uZ**2))
<br>
From there, we can observe several problems. Two examples:
<br>
1) (uX, uY, uZ) (uY, uX, 0)' = uX*uY + uY*uX =/= 0 (Most of the
time.)
<br>
2) || (uY, uX, 0) || =/= 1 if uZ =/= 0 (Since ||(uX, uY, uZ)|| =
1.)
<br>
Thus, the set of positions generated by that function does not
correspond exactly to the detector.
<br>
<br>
A possible correction: (it needs to be applied on the computation
of ap_Position1 and ap_Position2)
<br>
Replace (uY, uX, 0) by (uY, -uX, 0) / || (uY, -uX, 0) ||
<br>
Replace (uX*uZ, uY*uZ, sqrt(1-uZ**2)) by (-uX*uZ, -uY*uZ, 1-uZ**2)
/ || (-uX*uZ, -uY*uZ, 1-uZ**2) ||
<br>
<br>
Note: Since mp_crystalOrientationZ is 0.0 for all detectors in
most scanner, this error might have negligible repercussions in
practice. I am not sure, since I didn't test its repercussions in
reconstruction.
<br>
<br>
I have another question. In the comments prior to the function
GetRdmPositionsAndOrientations, there is the following line: "fix
the possibility to draw LOR outside the actual crystal volume (if
mp_meanDepthOfInteraction != 0.5)". From what I understand, that
function never uses that parameter. Was that comment meant for the
function GetPositionsAndOrientations or did I miss something?
<br>
<br>
Have a nice day,
<br>
Maxime Toussaint
<br>
<br>
_______________________________________________
<br>
Castor-users mailing list
<br>
<a class="moz-txt-link-abbreviated" href="mailto:Castor-users@lists.castor-project.org">Castor-users@lists.castor-project.org</a>
<br>
<a class="moz-txt-link-freetext" href="http://lists.castor-project.org/listinfo/castor-users">http://lists.castor-project.org/listinfo/castor-users</a>
<br>
</blockquote>
<br>