[Castor-users] Problem with sensitivity computed from normalization datafile

Simon Stute simon.stute at cea.fr
Mon Jul 23 16:27:22 CEST 2018


The normalization datafile should contain as many events as the number 
of possible lines-of-response that may record coincidences.

So at a certain point, you should start from the sinogram of 
normalization coefficients provided by Philips and create one 
"normalization event" per sinogram bin to build the normalization data 
file. You must know the corresponding crystals-pair for each sinogram 
bin and write it in the event.

So prompt coincidences have nothing to do with the sensitivity image 
computation here.

If you are using your list-mode events to build the normalization 
datafile, this is normal that you get an object in your sensitivity image.
It is simply as if you backprojected all your events.

Simon

Le 23/07/2018 à 15:50, Julien Salvadori a écrit :
> Thank you Simon for the quick reply.
>
> I disabled the attenuation and normalization in the calculation of the 
> sensitivity and i always get strictly the same result. To make sure, i 
> also reiterated the sensitivity computation by forcing the variable 
> lor_sensitivity to 1 in the ProcessThisLine function of 
> oSensitivityGenerator, with again the same result !
>
> Indeed,i use prompt coincidences to generate the normalization 
> datafile, it's not what is necessary to do ? and yes normalization 
> coefficients are provided by Philips.
>
> Julien
>
> 2018-07-23 10:40 GMT+02:00 Simon Stute <simon.stute at cea.fr 
> <mailto:simon.stute at cea.fr>>:
>
>     Hi Julien,
>
>     Looking at the SensiNorm, you are taking normalization and/or
>     attenuation into account, because we can see your object in the
>     sensitivity image.
>     If you really want to disable it, even if it is present in your
>     normalization data file, just use the "-ignore-corr attn,norm"
>     option, as you did in the reconstruction.
>     Otherwise, the attenuation is still present in the sensitivity
>     image, even if you ask for ignoring it in the reconstruction, as
>     you provide the sensitivity image.
>     I suggest you try this, until you get an almost similar image to
>     the one built with the scanner elements.
>
>     Also, when looking at the SensiNorm, it is quite noisy and we can
>     see the inserts which looks like as if you used the prompt data as
>     the normalization coefficients.
>     I am wondering what kind of normalization coefficients are you
>     using ? Are you using the one provided by Philips ?
>
>     Best
>     Simon
>
>
>
>
>
>     Le 22/07/2018 à 17:18, Julien Salvadori a écrit :
>>
>>     Dear Castor users,
>>
>>     I can reconstruct without problem a list mode from an IEC phantom
>>     recorded on the Philips Vereos TEP (with and without correction)
>>     when the sensitivity is calculated from the .geom scanner.
>>     However, when the sensitivity is generated from a normalization
>>     file, it does not work anymore. I have no idea where the problem
>>     may come from, especially since the sensitivity image generated
>>     from the normalization file seems correct to me.
>>
>>     Here in attachement :
>>
>>     “SensiScan” : Sensitivity calculated from the scanner element
>>     without taking into account attenuation and normalization.
>>
>>     “ReconWithSensiScan” : MLEM reconstruction with sensitivity
>>     “SensiScan” without any correction
>>
>>     èGets reconstructed images which seems to me coherent considering
>>     the absence of correction
>>
>>     “SensiNorm”: Sensitivity calculated from the normalization
>>     datafile without taking into account attenuation and
>>     normalization (attenuation and normalization factor set to 1)
>>
>>     ReconWithSensiNorm : MLEM reconstruction with sensitivity
>>     “SensiNorm” without any correction
>>
>>     èGets an almost uniform image with values very close to 0
>>
>>     Do you think that the sensitivity image calculated with the
>>     normalization file seems correct? Have you ever encountered a
>>     similar problem?
>>
>>     Any help will be welcome!
>>
>>     Kind regards
>>
>>     Julien Salvadori
>>
>>     Phd student, IADI
>>
>>
>>>>
>>     1.zip
>>     <https://drive.google.com/file/d/1oU9veGi2I5sE8fS0-aeutq0HsUeTFn-Y/view?usp=drive_web>
>>>>
>>
>>     _______________________________________________
>>     Castor-users mailing list
>>     Castor-users at lists.castor-project.org
>>     <mailto:Castor-users at lists.castor-project.org>
>>     http://lists.castor-project.org/listinfo/castor-users
>>     <http://lists.castor-project.org/listinfo/castor-users>
>
>
>     _______________________________________________
>     Castor-users mailing list
>     Castor-users at lists.castor-project.org
>     <mailto:Castor-users at lists.castor-project.org>
>     http://lists.castor-project.org/listinfo/castor-users
>     <http://lists.castor-project.org/listinfo/castor-users>
>
>
>
>
> _______________________________________________
> Castor-users mailing list
> Castor-users at lists.castor-project.org
> http://lists.castor-project.org/listinfo/castor-users

-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://lists.castor-project.org/pipermail/castor-users/attachments/20180723/3a8d22af/attachment-0001.html>


More information about the Castor-users mailing list