[Castor-users] Normalization

STUTE Simon simon.stute at chu-nantes.fr
Tue Sep 15 17:15:21 CEST 2020


Oh yes indeed, figure 5 of the reference provides a very clear demonstration of the effect !

Thanks for the interesting discussion.

Best,

Simon

________________________________
De : Sadek Nehmeh <san2028 at med.cornell.edu>
Envoyé : mardi 15 septembre 2020 15:28
À : STUTE Simon; castor-users at lists.castor-project.org
Objet : RE: Normalization

Yes it is.

Summing the images did not show any artifacts. As I mentioned in my previous message, I can see mild ring artifacts with 30min of data.
I used an energy threshold of 435 to 650 keV. My signal was integrated on the detector block level. I also have, transaxially, one crystal size gap between the rsectors (each rsector included 4x2 blocks), which unless a normalization correction is applied should create ring artifacs, but those were not obvious here.

For castor, I used the Joseph projector, and the images were reconstructed without filtering or PSF; I sent those off hoping to be able to see the rings.

One major difference from STIR though is that it (I mean STIR) assumes a cylindrical detector geometry, while castor (please correct me if I am wrong) uses the correct block geometry. This can have major effect on accounting for the image artifacts. This is nicely described in Khateri et al. https://ejnmmiphys.springeropen.com/articles/10.1186/s40658-019-0248-9

Anyway, all what I can say is that castor is doing pretty good job!

Regards,
Sadek

From: STUTE Simon <simon.stute at chu-nantes.fr>
Sent: Tuesday, September 15, 2020 9:08 AM
To: Sadek Nehmeh <san2028 at med.cornell.edu>; castor-users at lists.castor-project.org
Subject: [EXTERNAL] RE: Normalization


Indeed the image is good !

And if you sum all slices of the left image, do you see rings with the 2.5mn scan ?



Did you include an energy threshold in GATE ? If yes, then did you include a readout module, and a which level ? In simulations, the main effect creating rings is usually due to events scattering at boundaries of block detectors leading to a loss of efficiency in crystals at these boundaries.



As for the difference with STIR, I cannot tell that much. At least I can tell that in CASToR, the exact crystal positions are used for the reconstruction.

What projector did you use ? Also did you include an image-based PSF ? Because it can lessen the artifacts to some extent.



Simon



________________________________
De : Sadek Nehmeh <san2028 at med.cornell.edu<mailto:san2028 at med.cornell.edu>>
Envoyé : lundi 14 septembre 2020 19:29
À : STUTE Simon; castor-users at lists.castor-project.org<mailto:castor-users at lists.castor-project.org>
Objet : RE: Normalization

Many thanks Simon for your detailed response. Please see below an example transaxial slice and the summed coronal image for a 2.5 min of simulation. The images are artifact-free event though no normalization correction was applied. I had to go with 30min of data to see the ring artifacts, even though very mild especially when compared to I am used to from STIR.

Regards,
Sadek




[cid:image001.png at 01D68B40.FC605DB0]




From: STUTE Simon <simon.stute at chu-nantes.fr<mailto:simon.stute at chu-nantes.fr>>
Sent: Monday, September 14, 2020 5:26 AM
To: Sadek Nehmeh <san2028 at med.cornell.edu<mailto:san2028 at med.cornell.edu>>; castor-users at lists.castor-project.org<mailto:castor-users at lists.castor-project.org>
Subject: [EXTERNAL] RE: Normalization


Dear Sadek,



The PET sensitivity image is computed using one of the three following method :

 - if a histogram datafile is provided, then the sensitivity is computed by backward-projecting the product of attenuation and normalization corrections of each LOR of the datafile, one-the-fly for each subset

 - if a listmode datafile is provided, without normalization datafile, a double loop is performed over all crystals of the scanner so that for each possible crystal pair, the corresponding line joining the two crystals is used to forward-project the attenuation image (if provided) and backward-project the corresponding attenuation coefficient factor in the sensitivity image. In that case, no scanner normalization is taken into account so that it is implicitly assumed that all crystal pairs have the same efficiency

 - if a listmode datafile is provided, a normalization datafile can be provided so that the sensitivity image will be computed from it. This normalization datafile contains all LORs that may detect events, with associated attenuation and normalization coefficients. A loop is done over all these LORs and the coefficients are backward projected.



Could you send an image that you reconstructed ?



If you did not proceed to the simulation of a normalization scan (using a rotating source or a "donnut" source) in order to compute normalization coefficients and build a normalization datafile, then it is very likely that you have artifacts in your image.



Did you try to sum up all slices of the reconstructed image ? Do you have sufficient statistics in your dataset ?



Best

Simon



________________________________
De : Castor-users <castor-users-bounces at lists.castor-project.org<mailto:castor-users-bounces at lists.castor-project.org>> de la part de Sadek Nehmeh <san2028 at med.cornell.edu<mailto:san2028 at med.cornell.edu>>
Envoyé : vendredi 11 septembre 2020 18:26
À : castor-users at lists.castor-project.org<mailto:castor-users at lists.castor-project.org>
Objet : [Castor-users] Normalization

Dear Castor Developers and Users,

I have successfully reconstructed the GATE images for a Biograph Vision. The images look great and artifact-free. The interesting thing is that even though I did not apply any normalization correction, the images do not show any rings artifacts! Could someone please clarify some details about how the sensitivity image is produced and how the algorithm compensates for the expected ring artifacts?

Thanks,
Sadek
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://lists.castor-project.org/pipermail/castor-users/attachments/20200915/1f79d212/attachment-0001.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image001.png
Type: image/png
Size: 31311 bytes
Desc: image001.png
URL: <http://lists.castor-project.org/pipermail/castor-users/attachments/20200915/1f79d212/attachment-0001.png>


More information about the Castor-users mailing list