[Castor-users] Scatter rate estimation for LM-TOF-MLEM

jbaran jakub.baran at ifj.edu.pl
Mon Sep 20 13:18:22 CEST 2021


Hi Marina,

many thanks for your comments. Unfortunately, the solution you provided 
didn't change the results much. The greater background in the middle of 
the imaged object, characteristic for the lack of scatter correction is 
still visible. We are suspecting that probably we need greater 
coincidence statistics to sort it out.

I will let you know if we found anything interesting.

All the best,
Jakub

On 2021-08-26 09:08, Marina Filipovic wrote:
> Hi Jakub,
> 
> You should divide your scatter rate with the size of the TOF bin
> converted to mm. That's if you're using the "continuous" LM-TOF recon,
> which is run if you don't specify the size of the TOF bin in the
> header. If you specify the size of the TOF bin in the header, then you
> don't have to change your scatter rates per bin, that's easier, I
> should have mentioned this option first. The TOF resolution field is
> only for TOF resolution. The documentation is maybe not so clear about
> these options?
> Let me know if this helps.
> 
> Cheers
> 
> Marina
> 
> On Wed, 25 Aug 2021, 08:52 jbaran, <jakub.baran at ifj.edu.pl> wrote:
> 
>> Hi Marina,
>> 
>> many thanks for your answer. could you please clarify what do you
>> mean
>> to "divide the estimated TOF bin scatter rates with the spatial
>> width of
>> the TOF bin, directly in the list-mode datafile". Once I estimate
>> the
>> scatter fraction I presume I should divide it by the TOF bin (e.g.
>> 200ps)? However, I do not specify the TOF bin size in my header.
>> Unless
>> the bin size is then defined as a TOF resolution?
>> 
>> Example of the header which I am using:
>> 
>> Data filename: data_df.Cdf
>> Number of events: 298175691
>> Data mode: list-mode
>> Data type: PET
>> Start time (s): 0
>> Duration (s): 500
>> Scanner name: scanner1
>> Calibration factor: 1
>> Isotope: unknown
>> TOF information flag: 1
>> TOF resolution (ps): 200
>> List TOF measurement range (ps): 3000
>> Scatter correction flag: 1
>> 
>> What is the way to proceed in that case?
>> 
>> All the best,
>> Jakub
>> 
>> On 2021-08-18 16:54, Marina Filipovic wrote:
>>> Hi Jacub,
>>> 
>>> Have you divided the estimated TOF bin scatter rates with the
>> spatial
>>> width of the TOF bin, directly in the list-mode datafile, as
>> explained
>>> in the documentation (general and TOF-specific) ?
>>> 
>>> Marina
>>> 
>>> On Wed, 18 Aug 2021, 00:52 jbaran, <jakub.baran at ifj.edu.pl> wrote:
>>> 
>>>> Dear Castor Developers and Users,
>>>> 
>>>> I am trying to incorporate the information about the scatter
>>>> correction
>>>> in LM-TOF-MLEM reconstruction. I have calculated the scatter
>>>> fraction
>>>> for each TOF bin. The values of scatter rate are in range 1-2. I
>>>> have
>>>> applied them for the reconstruction, however the reconstructed
>>>> images
>>>> seem to be not corrected for scatter. Please find the attached
>>>> screenshot. On the top panel, the reconstructed image with
>>>> true+scatter
>>>> coincidences is presented and on the bottom the same coincidences
>>>> were
>>>> used but with scatter correction applied. On the right
>> corresponding
>>>> 
>>>> line profiles are presented (ots - no scatter correction;
>> corrected1
>>>> -
>>>> scatter correction applied)
>>>> 
>>>> FYI: The total number of coincidences used for the reconstruction
>> is
>>>> 1
>>>> order of magnitude lower than the number of possible LORs.
>>>> 
>>>> Do you have any idea what kind of values should be applied for
>> the
>>>> scatter rate? Did you have any similar case and found the
>> solution
>>>> how
>>>> the scatter rate should be calculated?
>>>> 
>>>> All the best,
>>>> Jakub
>>>> _______________________________________________
>>>> Castor-users mailing list
>>>> Castor-users at lists.castor-project.org
>>>> https://lists.castor-project.org/mailman/listinfo/castor-users
>>>> 
>>>> To look for something in the mailing-list archives, use the
>> search
>>>> box at:
>>>> https://castor-project.org/mailing-list


More information about the Castor-users mailing list