[Castor-users] Scatter rate estimation for LM-TOF-MLEM

jbaran jbaran at ifj.edu.pl
Tue Dec 7 10:59:51 CET 2021


Hi Simon,

many thanks for the quick response. The answers are as follows:


On 2021-12-07 09:56, STUTE Simon wrote:
> Hello Jakub,
> 
> Thanks for the detailed email !
> 
> Do you take the scan duration into account in the datafile header ?

Yes, I added the information about the scan duration in the header. On 
the other hand in the reconstruction I used -ignore-corr fdur option.

> 
> You say that you estimated the scatter rate for each TOF bin
> independently to be 1-2. Is it "1 or 2" (which would be insanely high)
> or "e-2" ?

I was not explicit here. The scatter rate passed to the castor lm is 
ranging from 0 to 2. However most values are in the range from 1e-3 to 
1e-1.

> 
> How did you compute this estimated scatter rate ?

We made basic estimation based on Monte Carlo simulations - NEMA IEC 
study. For the cubic study we use -sc flag during the conversion of the 
ROOT data to CASToR. For this the scatter rate is at the level of  about 
1e-4.

> 
> And how did you translate this information into your list-mode file ?

We adapt the CASToR code to read the scatter correction rate factor from 
the matrix provided by ourselves.

> 
> Best
> Simon

All the best,
Jakub